Quote:
Originally Posted by ericst11
Jerry i'm testing right now . I had a small tube of agar That a friend gave me from doing some flask for me and i am testing with that . I sterilized with rubbing alcohol and let it run for about 30 min and then sterilized the tube and un caped it hopefully it is contaminit free. By the way i have a minihelic meter it reads at 1.0 is that normal. Thanks Jerry
|
I am unsure just what medium was in the tube. Is it the same medium you used to sow your seed on? If so, I would not necessarily depend upon that medium for filter integrity checks. I would use a good general bacteriological culture media and, as mentioned, another one that is designed for culture of fungi. Also, what is the diameter of the opening on the tube. The most common opening sizes of test tubes are (or used to be) 13, 16, and 25 mm. That size opening is too small to use for testing. Most ready prepared media comes in 100 mm plates thus exposing a much larger surface area.
I don't know what type gauge you have so I can't comment on your reading of 1.0 (and to be truthful, my use of these type gauges happened a lifetime ago). The gauges we used specifically on laminar flow hoods were used to measure the pressure drop across the HEPA. If I remember correctly, a normal reading was usually somewhere between O.25 and 0.50. Higher readings usually meant that the HEPA was ready for replacement (the filters usually lasted 4-6 years with heavy use, but in a clean environment).
If the cost in not prohibitive to you, I think I would still opt to purchase some plates of prepared media as mentioned and proceed from there.