At 125 ppm N, your stock solution
should have around 288 ppm from the N, P, & K alone.
I can think of three simple possibilities: 1) That TDS meter is not calibrated correctly (it really takes ate least 2 data points to establish the correct calibration curve), 2) by adding Pro-Tekt to your stock solution, you have precipitated come of the fertilizer constituents (Pro-Tekt and fertilizer should not be mixed until they are in their final dilution), or 3) both.